I have a custom data acquisition board/controller with a Cyclone 4 (EP4C15) and a pair of ADAS3022 ADCs running in parallel. Both ADCs are being operated at 1MSps, with an effective rate of 125kSps per channel. The channels are scanned sequentially by the driver, with the data being dumped into a dual-port RAM where the channel number forms part of the address.  

 

That part appears to be working flawlessly, and I can use SignalTap to recreate perfect waveforms by tying into the write side of the memory interface between the driver and the controller. I checked several channels and the signals are being replicated just as they appear on the scope. (albeit with slightly lower resolution) 

 

Now, where things get hinky is on the read side. The controller reads some parameters at 100 Hz, and other parameters at 25 kHz. For the most part, the 100 Hz process works fine - it pulls four samples from the buffer, computes the differential voltages, does some error calculations and adjusts the system, then repeats until the differential error is within tolerable limits. This has been highly reliable so far, with no known instances of failure. 

 

The 25kHz process, on the other hand, has been having issues. Some of the signals we are measuring are nearly square, and when the data is being pushed out for collection, we see a lot of junk in the data. Specifically, repeated edges, dropouts, etc. I initially thought that perhaps it was a memory access problem, so I rewrote the controller to bypass the DMA mechanism and literally copy/paste using CPU instructions, only to see identical garbage data. Now I am beginning to think there is some more fundamental issue. 

 

Now, my original thought was that I could subsample the output of the ADC driver, which is running at 125 kSps/channel, and the worse that would happen is that I might miss fast edges (which is fine, we are measuring fairly long square wave like signals) or be off a bit in time. Instead, it looks like random noise is being generated during every step function of the signal being measured. Once the signal is stable, this effect seems to go away, which leads me to conclude that perhaps this is an aliasing issue. 

 

As an aside, the memory interface between the ADC firmware and the controller CPU is double-buffered so that the ADC process isn't ever writing to the same page that the controller is reading. The ADC firmware verifies that the controller isn't reading it's side before flipping the page. Also, both sides are running at the same clock frequency (100MHz) and the entire design is making timing. I verified this by setting up a SDC file and running TimeQuest. This doesn't appear to be a low-level timing issue or RAM setup & hold issue. 

 

My current plan is to simply pace the ADC subsystem with the 25kHz process, rather than letting it collect samples at the ADC's fastest data rate, which should resolve the issue. However, I'm still not sure why my original design is having this issue, and I'd like to understand the underlying problem. 

 

Thanks!
0 Kudos
3 Replies
Altera_Forum
Honored Contributor II
682 Views

Hi, 

 

I can't completely follow what is going on but I think I get it. It does sound like you maybe see aliasing. The sampling process doesn't care if you sample at too low of a rate for the waveform at the input or skip samples coming out of a fast ADC, the impact can be similar. If the processes aren't completely synchronous then it might be hard to track down. If you were to downsample by filtering and decimating, then you can pull off skipping samples deliberately. 

 

If the signals are truly square and you see a lot of samples at "in-between" values, that is a little suspicious. 

 

I think if you set the sampling rates (and anti-aliasing filters) appropriate for the waveforms and always sample/process contiguous sample segments, you should be OK. As I said, I'm not completely sure I understand the assumptions. 

 

If the scenarios (waveforms and schedule of how samples are drawn) are easy to define, mocking up the sampling system in Matlab/Octave or your tool of choice might be a good approach if you think there might be more to the problem. 

 

Lance
0 Kudos
Altera_Forum
Honored Contributor II
682 Views

Lance, 

Thanks for the reply. As it turns out, the issue was two unrelated "problems". The first is a real bug, and we haven't found the source yet. The driver and the DMA system seem to occasionally get out of sync on power up. This is usually resolved by power cycling the system, but for some reason the software guy didn't think of that. A serious problem, but one that happens infrequently and is resolved easily. 

 

The other issue is that I forgot how the bus interface code worked. My board is driving a bridge to the host PC over a synchronous LVDS interface. Because my board is the clock source, I align the data on the falling edges of the clock. Due to pipelining of the block memories, I have to go through a few delay registers to line everything up properly, which is what was confusing me. Once I started qualifying on the clock correctly, everything was shown to be working at the output of the board. I just hadn't documented it well enough to remember a year and half later...